My pleasure!Hi janoshik - fellow chemist here (I work in synthetic chemistry, not analytical.). Just made an account to respond lol.
When you say a new testing method - is this a different solvent blend/column packing material? From the HPLC, it looks like you're getting much better separation of some other product (the 5.276 / 6.103 retention time products).
Any idea what those peaks are? Some sort of polymerization product or something?
Anywho, very cool!
This is different stationary phase and different gradient program - so both parameters you've mentioned changed.
If you see a bit of a non-Gaussian shape on the right shoulder of the GH in the old method (around 4.05 min), that's the 5.276 peak in the new method - going from no separation to a great one! I haven't done my own research on the identity of those impurities, but Karlsson made a few really nice publications about it and the stuff I'm seeing is the same he saw during his analyses - usually just very minor modifications of the GH molecule itself.
Also, if you check his article linked below, even he didn't achieve peak shape that was perfectly Gaussian, so one can only assume that I've chromatographically separated what he did not - if I may boast a little.
Ref: Separation of Recombinant Human Growth Hormone Variants by UHPLC
Polymerization is tested in the dimer and related proteins test - which is conducted via size exclusion chromatography, as RP-HPLC ain't quite cut for that.
Cheers